Quarter wave plate and specular reflections
boomblurt
Post by boomblurt »
I have a SMZ-U stereo microscope and am considering getting an epi illuminator. After lots of googling I am still confused as to how a quarter wave plate reduces specular reflections when using coaxial episcopic illumination.
Can anyone help me understand please?
Geoff
g4lab
Post by g4lab »
http://ipl.physics.harvard.edu/wp-uploa ... _s07_7.pdf
https://en.wikipedia.org/wiki/Polarizer ... polarizers
http://www.u-oplaz.com/new/waveplate.htm
https://pelilab.partners.org/papers/JAO ... rcular.pdf
The above one has a good diagram showing how it works to suppress glare on crt monitors. It works the same way on the incident illuminator. The light is also polarized.
The techniques don't work on metal or other electrically conductive surfaces.
Almost all coaxial illuminators come with a quarter wave plate attachment. The Zeiss version for their SV6/11 stereo has the quarter wave plate mounted in a plane that is non parallel to the front of the objective. There are also techniques included to reduce the glare from optics and internal components. A friend of mine got a CoAx illuminator for his M400 and thought it was worthless until I told him to find some quarter wave plate and apply it to the front of the objective. He could then see into gemstones with the coax illuminator. I have illuminators like that for the Zeiss SV11 , the Wild M400 and M series and Nikon SMZ10 they all came with quarter wave plates. As do zeiss Antiflex objectives. for compound scopes (incident light scopes) The quarter wave plate needs to be in a rotatable mount.
[Admin edit, 9/7/2000, RJL, to replace broken link for .../papers/JAOA%201986%20Circular.pdf]
Zeiss, who seems invented the system, calls it Antiflex
It serves to avoid reflections in dark subjects and to eliminate internal reflections of lens elements surfaces
From Microscopy from the very beginning Zeiss booklet, page 32
https://www.google.ca/url?sa=t&rct=j&q= ... 3917,d.d24
When objectives of a very low magnification are used, a
so-called Antiflex cap (a rotatable ?/4-Platte in front of the
objective) permits otherwise unavoidable reflection to be also
eliminated from “dark” sample surfaces.
See also Handbook of Incident Light Microscopy, page 39:
I only have used it (but few times, so little experience) with low power microscope objectives. Some high end stereos and macroscopes like the Nikon AZ use it.
boomblurt
Post by boomblurt »
So if I understand correctly the emitted light is polarised before it is reflected down and passes through the objective, and then goes through a crossed analyzer as it goes back up, past the illuminator to the eyepiece. The twofold pass through the 1/4 wave plate, located between the objective and specimen, rotates the direction of polarization so it can pass the analyzer, while other (internal reflection) light is blocked?
So ... the purpose is to reduce internal reflections - not glare off the specimen? No wonder I was confused. :P
EDIT : That seems to be, I think, the idea of the Antiflex system (Pau's link, page 39). But I think I need to chew this over some more, in view of Gene's last link that show reduction of specimen reflections. Thanks for the links.
Geoff
g4lab
Post by g4lab »
The reflection it supresses is on the specimen. It is only internal reflection if the "specimen" is the surface of an old CRT monitor behind one of these screens that work on the very same principle.
boomblurt
Post by boomblurt »
Gene, is the diagram on your last link at page 299 the right idea - normally reflected light reverses its rotation of polarity and therefore is not passed back through the linear polarizer (after the 1/4 wave retarder)?
EDIT: With a crude setup of a torch (flashlight), linear polarizer and piece of cellophane as a workable 1/4 wave retarder I am able to see an effect of reduced glare that varied with rotation of the cellophane. (The cellophane was tested as a wave retarder by looking at polarized light - a laptop LED screen - with the polarizer between it and me. With the polarizer set for maximum extinction the cellophane was rotated and light varied as expected - four min/max transmissions per rotation. I had a tried a normal photography CPL filter but something was going on there as light could only be extinguished with the polarizer closest to the screen).
Geoff
boomblurt wrote:I had a tried a normal photography CPL filter but something was going on there as light could only be extinguished with the polarizer closest to the screen).
Here is a reliable demonstration.
Take the CPL filter to an ordinary mirror. Reverse the filter so that its normal front is facing your eye and its normal rear (with external threads) is toward the mirror. Using only one eye, look through the filter at its reflection in the mirror. The filter will appear opaque black -- you will not be able to see your own eye although you will still be able to see other parts of your face around the filter. This is because light going out through the filter becomes circularly polarized, the mirror reverses the direction of polarization, and the filter then blocks that light on its return.
It works the same way in a microscope. The combination of linear polarizer and 1/4 wave retarder causes the light to be circularly polarized, anything that acts like a mirror surface reverses the polarization, and the retarder+polarizer blocks that light on return. The bit about "acts like a mirror surface" includes any lens surface, specular reflections from the specimen, and reflections from most types of metal. It does not include subsurface reflections from organic material.
When you're demonstrating with a CPL, the effect depends on orientation of the filter. If you do not reverse the filter, leaving its normal rear with external threads facing your eye, then it will not go opaque black, only a little dim. This is because failing to reverse the filter sets up an optical chain that is retarder/polarizer/mirror/polarizer/retarder. The mirror is then working with linear polarized light, whose direction it does not affect, so the polarizers are effectively parallel and not crossed. To get the darkening, you need to reverse the filter so that the chain is polarizer/retarder/mirror/retarder/polarizer, so that the mirror can work on the circular polarized light to reverse its direction.
The 1/4 wave plate doesn't eliminate subject reflections but allows to control them like rotating one of the polarizers to intermediate positions between parallel and crossed. It really allows to better contrast than plain bright field.
To really suppress specular reflections on subject what you need is cross polarization. I do it in my stereo without Antiflex with a ring polarizer over the ring fiber optic illuminator or with polarizers on the tips of fiber optics bundles (no actual epi).
Take a look at http://www.photomacrography.net/forum/v ... p?p=154251 and follow the links in it.
From http://zeiss-campus.magnet.fsu.edu/arti ... ected.html
In cases where objectives of very low magnification are used in reflected polarized light, a rotatable optical plate (termed an Antiflex cap) consisting of a one-quarter wavelength lambda plate is placed on the objective front lens element to block reflections from the objective itself. The Antiflex method is also particularly useful when the specimen has very low reflectivity, such as would be observed in coal samples.
JH
Post by JH »
Interesting discussion!
If it is of any interest I used an ordinary cirkular polarizing camera filter between the ant and the objective in the third picture in this post
http://www.photomacrography.net/forum/v ... hp?t=22034
Regards Jörgen
boomblurt
Post by boomblurt »
Thanks all for your information and patience. Rik's explanation (same idea as the diagram in Gene's last link, p299) makes sense. I haven't used any of this equipment so am trying to understand purely on the basis of (my very limited) theory, but things seem to be a bit clearer now - hopefully.
Most of the links I had seen before - eg the Zeiss handbook was linked on one of the Yahoo groups but that mentions a crossed polarizer/analyzer. This seems to be a different setup to a single polarizer+retarder and although the principles are the same it confused me. I haven't been able to find any specifics of the SMZ-U coaxial illuminator.
Regards my LED monitor 'tests' I had simply(!) assumed that the photographic CPL was composed of a separate polarizer and wave plate that rotated independently of each other (as in my cellophane combo). This would be a very bad way to design a CPL filter. As a side note, the cellophane worked better than I expected.
Rik's mirror experiment was interesting and seemed to work well with a CPL filter. But when I tried it with the analyzer from my compound microscope (~1/2" diameter) which presumably is also circularly polarized (as it will only cross polarize in one direction) the effect was smaller. Could this be due to the amount of reflections between the mirror and the polarizer, or something else?
Geoff
boomblurt wrote:Rik's mirror experiment was interesting and seemed to work well with a CPL filter. But when I tried it with the analyzer from my compound microscope (~1/2" diameter) which presumably is also circularly polarized (as it will only cross polarize in one direction) the effect was smaller. Could this be due to the amount of reflections between the mirror and the polarizer, or something else?
Given the behavior of only cross polarizing in one direction, and darkening some but not completely in the mirror test, I would guess that the analyzer is an elliptical polarizer. See https://www.newport.com/images/webDocum ... /18954.pdf for discussion.
The technique that Pau describes, rotating the 1/4 wave plate to control the reflections, is based on elliptical polarization.
boomblurt
Post by boomblurt »
Aaahh .. thanks Rik!
Geoff
blekenbleu
Re: CPL and mirrors
Post by blekenbleu »
boomblurt wrote:
Sat Feb 20, 2016 8:49 pm
First, apologies if replying to very old topics is uncool, but:
1) Gene's last link, p299 needs an updated URL: https://pelilab.partners.org/papers/JAO ... rcular.pdf
2) Thinking about circular polarized reflection in mirror as twice 1/4 wave rotation gets confusing,
when repeating the experiment with one and then two circular polarizers:
* view of second circular polarizer thru the first maintains extinction, independent of rotating either.
This implies that extinction results from mirror reversing circular polarization handedness,
then mirrored circular polarizers of the same handedness will mutually extinguish.
All (camera) circular polarizers tested so far evidence the same handedness.
Having just been acquired from https://www.ebay.com/str/techno2020,
a quarter wave plate was inserted between circular polarizer and the mirror.
This cancelled extinction, since two passes thru quarter wave plate = half wave plate,
which reverses circular polarization handedness, cancelling mirror handedness reversal.
Carefully aligning things allows viewing semi-infinite recursive handedness reversals.
Rotating the quarter wave plate relative to the circular polarizer made little difference viewed in the mirror,
unlike substantial differences when rotating that quarter wave plate between a linear or circular polarizer and a polarized light source (LCD monitor),
seemingly approximating a de Sénarmont compensator.
Re: CPL and mirrors
blekenbleu wrote:
Wed Sep 07, 2022 9:12 am
Done, thanks for the heads-up.
This is true in a formal sense, but personally I think I get more clarity from going one step deeper in the details.
The initial light path consists of [LP->QWP] --> mirror --> [QWP->LP], where LP means a linear polarizer and QWP is a quarter wave plate, with the pair oriented to form a circular polarizer. In this configuration, the starting pair makes the light be circularly polarized, the mirror reverses the handedness, and the resulting circularly polarized light is blocked by its return path through the same pair.
With the addition of another QWP, the configuration becomes [LP->QWP] --> QWP --> mirror --> QWP --> [QWP->LP], with no particular orientation between the two QWP's. In this configuration the passage of circularly polarized light through the added QWP converts the light back to linearly polarized, with its linear orientation matching the orientation of the added QWP. There is zero handedness for the mirror to reverse. Instead, the mirror simply preserves the linear orientation, which is automatically aligned so that passage back through the QWP will convert it back to circular with the proper handedness to get back through the final circular polarizing pair.
Note that replacing the single mirror with a combination of slanted mirrors can change the orientation of the linear polarization. I have not worked through this case in enough detail to be sure, but I think that changing the polarization angle can play in interesting ways with the plates on the return trip. If you are interested in exploring this aspect, I suggest to start with a collection of chrome balls such as illustrated at https://www.photomacrography.net/forum/ ... hp?t=33842 "Cross-polarization and multi-bounce specular reflections".
blekenbleu
Re: CPL and mirrors
rjlittlefield wrote:
Wed Sep 07, 2022 3:20 pm
A wall mirror and second hinged mirror to its left
made observing a second reflection cancelling circular polarization extinction relatively easy:
Increasing depth of focus to get everything sharp eluded me.
Oops, thanks for that teaching; linearly polarized single reflection of QWP between CPL and mirror
was confirmed by viewing that reflection thru a linear polarizer.
I read that and am very interested in managing epi microscopy specular reflections
but lack both shiny balls and agility sorting what reflects from where; 2 flat angled mirrors sufficiently confuse.
Experimenting confirmed linear polarizer second reflections being more or less extinguished depending on orientation.
Seemingly, one wants adjustable orientations for epi microscope polarizer, analyzer and quarter wave plate...?