Based on a post by Scarodactyl the urge to dig out an old, and due to formatting problems never posted, text was rekindled.
For compound microscopes the numerical aperture (NA) is given exclusively by the objective. For zoom instruments it depends also on the zoom range and the entrance pupil of the zoom system. This is only rarely specified from the manufactures; indeed, the only tabulated specifications I have been able to find is for the Nikon HR Plan Apo 1x lens for the Nikon SMZ1500 stereo microscope (http://www.microscopyu.com/articles/ste ... intro.html) and as supplementary material in an article for the Olympus MVX10 with the MVPLAPO 2xc objective. Scarodactyl has provided data for the Leica Z6 APO and Z16 APO macroscopes and the MZ16/MZ16A stereo microscopes.
In general, the NA of zoom microscopes depends on the dimensions of the Common Main Objective (CMO), as well as on the dimensions of the zoom system. This is what Leica has elegantly implemented in their FusionOptics microscope, combining two optical zoom paths of different dimensions, emphasizing resolution and DOF respectively.
In the table below I have used the information on the Nikon HR Plan Apo 1x, together with the very useful spreadsheet on camera resolution and relay optics by Charles Krebs. The resolution is calculated as the number of pixels needed to cover the diagonal (22, 23 or 24mm, depending on the instrument), using 2 pixel/detail - in practice one may choose a higher pixel/detail ratio.
Code: Select all
Zoom factor NA Pixels/diagonal
---------------------------------------------------------------------------------------
Nikon SMZ1500 0.75 0.023 2413
HR Plan Apo 1x 1 0.029 3043
2 0.052 2727
4 0.085 2230
6 0.104 1819
8 0.118 1548
10 0.128 1343
11.25 0.131 1222
For the Olympus MVX10, subjected to the same calculatory treatment, one can see that the best performance is in the middle magnification region - something that may well also be true for other macroscopes. Conceivably this is a characteristic of the zoom magnification system rather than the objective. For the MVX10 the sweet spot is from 1-4 times zoom factor.
Code: Select all
Zoom factor NA Pixels/diagonal
-----------------------------------------------------------------------------------------
Olympus MVX10 0.63 0.070 5829
MVPLAPO 2xc 0.80 0.070 4590
1.00 0.110 5770
1.25 0.140 5875
1.60 0.190 6230
2.00 0.230 6033
2.50 0.295 6190
3.20 0.380 6230
4.00 0.475 6230
5.00 0.495 5193
6.30 0.495 4122
Code: Select all
Zoom factor NA Pixels/diagonal
-----------------------------------------------------------------------------------------
Leica MZ16/MZ16A 0.71 0.084 2962
Planapo 5x/0.5 1.00 0.110 2754
(10447243) 1.60 0.170 2660
Factor 4x 2.00 0.200 2503
FN21 2.50 0.240 2404
3.20 0.300 2347
4.00 0.340 2128
5.00 0.400 2003
6.30 0.450 1788
8.00 0.490 1534
10.00 0.500 1252
11.50 0.500 1089
That is not a function of the objective is supported by the data that the Leica MacroFluo 6.3:1 or 16:1 (the same as the Leica Z6 APO (6.3:1 Zoom) or Z16 APO (16:1 Zoom)) - apparently does not vignette with the eyepieces HC Plan 10×/25. (11507808) when used with the Planapo 5x/0.5 (10447243).
Below is a table encompassing some of the current macroscopes for which data can be readily found. This table now includes the Leica Z6 APO and Z16 APO instruments, where specifications were kindly provided by Scarodactyl. Data are for the instruments used at the highest possible magnification.
It should be noted, that Zeiss in at least one place have claimed an NA of 0.50 with a FOV of 1.5 mm for the PlanNeoFluar Z 2.3x, however, a manual for the instrument has not been found, and consequently this entry into the table should perhaps be treated with care. If the information is correct, an NA of 0.50 is seen with a zoom factor of 6.67, yielding a total magnification of 15.18 (**). That the highest zoom-factor is not conductive to the highest possible pixel count may also be why the Nikon AZ100 fares comparatively poorly in this exercise.
For comparison is included some of the often used Mitutoyo objectives as well as the (rather expensive) Mitutoyo HR Plan Apo 5x and 10x, which fares exceedingly well in this comparison. According to Mitutoyo literature, these objectives covers an image circle of 30mm (in practice for some more, ***).
Code: Select all
Objective Zoom factor NA Pixels/diagonal WD FOV@10x ocular
-----------------------------------------------------------------------------------------------------------------------
Nikon AZ100:
AZ Plan Apo 0.5x 8 0.05 1311 54 mm 5.50 mm (22mm)
AZ Plan Apo 1x 8 0.10 1311 35 mm 2.80 mm (22mm)
AZ Plan Fluor 2x 8 0.20 1311 45 mm 1.40 mm (22mm)
AZ Plan Apo 4x 8 0.40 1311 20 mm 0.70 mm (22mm)
AZ Plan Fluor 5x 8 0.50 1311 15 mm 0.55 mm (22mm)
Olympus MVX10
MVPLAPO 0.63x 6.3 0.15 3965 87 mm 5.54 mm (22mm)
MVPLAPO 1x 6.3 0.25 4163 65 mm 3.49 mm (22mm)
MVPLAPO 2xc 6.3 0.50 4163 20 mm 1.74 mm (22mm)
Leica Z6
Planapochromat 1x 4.5 0.117 3100 97 mm 5.10 mm (25mm)
Planapochromat 2x 4.5 0.234 3100 39 mm 2.68 mm (25mm)
Leica Z16
Planapochromat 1x 11.5 0.112 1161 97 mm 2.00 mm (25mm)
Planapochromat 2x 11.5 0.224 1161 39 mm 1.04 mm (25mm)
Zeiss Axio Zoom V16
PlanApo Z 0.5x 11.2 0.125 2448 114 mm 4.10 mm (23mm)
PlanNeoFluar Z 1x 11.2 0.25 2448 56 mm 2.00 mm (23mm)
PlanApo Z 1x 11.2 0.25 2448 60 mm 2.00 mm (23mm)
Apo Z 1.5x 11.2 0.37 2416 30 mm 1.40 mm (23mm)
PlanNeoFluar Z 2.3x 11.2 0.57 2427 10.6 mm 0.90 mm (23mm)
PlanNeoFluar Z 2.3x ** 6.67 0.50 3575 10.6 mm 1.50 mm (23mm)
Mitutoyo FS70 ***
Plan Apo HR 5x 1 0.21 4807 25.5 mm 4.80 mm (24mm)
Plan Apo HR 10x 1 0.42 4807 15 mm 2.40 mm (24mm)
Plan Apo 5x 1 0.14 3205 34 mm 4.80 mm (24mm)
Plan Apo 7.5x 1 0.21 3205 34 mm 3.60 mm (24mm)
Plan Apo 10x 1 0.28 3205 34 mm 2.40 mm (24mm)
Plan Apo 20x 1 0.42 2404 20 mm 1.20 mm (24mm)
From the above table can also be seen that a macroscope such as the Nikon AZ100 has a very consistent, if not particularly impressive performance. A slightly more modern instrument, the Olympus MVX10 macroscope is likewise consistent in performance, with a vastly better resolving power. In all the zoom instruments above, the front objectives are of the infinity type and the zoom system enlarges a portion of the "image" projected by the front objective.
One intriguing question therefore is to what extent the front objectives from macroscopes are good over larger areas; i.e. if a front objective can be directly coupled to an ordinary large aperture tele-lens in the role of a tube lens. Incidentally, the preceding should also hold true for front objectives of Common Main Objective (CMO) stereomicroscopes of the infinity type.
Placing an infinity CMO stereo microscope objective in front of a medium tele-lens (135mm f2.0) seems to indicate that this will work. The threaded part of the CMO objective was in this, highly unscientific, test made to rest on the UV-filter while the 135mm lens was focused at infinity. It is currently not known to me if optical parts of the zoom system works as correcting tube lenses, however, as it is a zoom system I am inclined to think that all corrections are made in the objectives.
Looking into using CMO objectives in homebuilt systems may prove a worthwhile exercise.