My Meopta objectives have poor contrast. Resolution seems to be ok? but it is like looking through a mist !
Do they 'just' need cleaning. If so is there a good "How to" ? Not sure of its vintage, guess at 60s or earlier, would acetone be ok ?
The 1st pic is of the Meopta 10:1 and the 2nd is another10:1 that I am using as a reference.
The 3rd pic is with the Meopta 45:1 and is dismal as you can see when compared with the ref10 scaled up by 430%, in the 4th pic.
The three 'pixel peek' comparisons show that the resolution of the 45 is, as would be hoped, better than the scaled-up 10.
Thanks for any thoughts?
The specimen is a dried blood smear. My blood ! peeling potatoes is dangerous and should be left to experienced significant_others !!
Meopta lens problems
Moderators: rjlittlefield, ChrisR, Chris S., Pau
-
- Posts: 1632
- Joined: Sat Apr 14, 2018 10:26 am
Re: Meopta lens problems
They could be grimy. There could be internal mold or delamination. You'd need to have a close look at them. A stereo microscope works well for this type of inspection.
-
- Posts: 1152
- Joined: Thu Mar 07, 2013 5:24 am
Re: Meopta lens problems
Hi,
The 45:1 requires the object to be covered with a cover glass. Ideally, there should also be a mountant between the blood smear and the cover glass as well (balsam or a drop of immersion oil). The open, uncovered smear results in strong spherical aberration. https://www.olympus-lifescience.com/fr/ ... spherical/
Never, ever use acetone on these vintage optics (the very worst). Some older systems don't even tolerate ethanol. Xylene was recommended by most companies and medical grade petroleum ether is a good and safe choice. There are some excellent instructions here: https://microscopy.duke.edu/sites/micro ... scsope.pdf
The 45:1 requires the object to be covered with a cover glass. Ideally, there should also be a mountant between the blood smear and the cover glass as well (balsam or a drop of immersion oil). The open, uncovered smear results in strong spherical aberration. https://www.olympus-lifescience.com/fr/ ... spherical/
Never, ever use acetone on these vintage optics (the very worst). Some older systems don't even tolerate ethanol. Xylene was recommended by most companies and medical grade petroleum ether is a good and safe choice. There are some excellent instructions here: https://microscopy.duke.edu/sites/micro ... scsope.pdf
Re: Meopta lens problems
Thank you both,
mold and delamination sound scarily terminal !
and I promise never ever to let acetone anywhere near them. I suspected that may be contraindicated, hence the question but that is all that I have at the moment except household detergent ! so will give them a reprieve for now.
Thanks for the recommendation to the pdf.
Immersion oil, I wouldn't have thought of that, I have several small bottles unused, in my box of inherited gifts. No other mountants came my way. I have been using the Legg-Dioni fructose methods. I also learned from OliverK that I should have diluted it with water, oh well, next time I stab myself many other experiments will follow
mold and delamination sound scarily terminal !
and I promise never ever to let acetone anywhere near them. I suspected that may be contraindicated, hence the question but that is all that I have at the moment except household detergent ! so will give them a reprieve for now.
Thanks for the recommendation to the pdf.
Immersion oil, I wouldn't have thought of that, I have several small bottles unused, in my box of inherited gifts. No other mountants came my way. I have been using the Legg-Dioni fructose methods. I also learned from OliverK that I should have diluted it with water, oh well, next time I stab myself many other experiments will follow
Re: Meopta lens problems
Previously, a 3: 1 mixture of diethyl ether and ethanol was used in Meopta to clean optics. Acetone + ethanol of about 9: 1 is now used.
Re: Meopta lens problems
Thank you for the information.
To be sure I am understanding correctly : they now use (Acetone+ethanol 9: 1) on both old and new lenses ?
Re: Meopta lens problems
Meopta service of old lenses no longer does ...
Acetone is only used for current production.
Acetone is only used for current production.