I've been playing around with the Leuchtbild condenser...
This time the goal was to get darkfield with a 63/1.4 oil immersion lens without stopping this lens down.
Placing a 14 mm stop below the condenser was the key to do so...
The diatoms were embedded in Naphrax, found at the Hagenburger Kanal, Schaumburg, Germany.
The pictures:
Detail:
Central stop in offset position:
finally one more A.P.:
Technikque: Zeiss Standard 18; Planapo 63/1.4Öl; Leuchtbild condenser;
S-Kpl 16x/12,5 and Kpl 20x/8; relay lens f=63mm; Canon EOS 5DII.
Best wishes
Jens
Diatoms in high nA DF.
- myriophyllum
Very nice work, Jens!
Your Leuchtbild condenser is an 1.4-1.6 NA one, correct? So centering a 14mm opaque stop under the condenser was all that was required for a full NA 1.4 objective to work?
I wonder if I can do something similar, with a 1.2-1.4 DF condenser, for a full NA 1.2 objective.
Thank you for sharing!
PS: I think you will resolve even more dots, if you image diatoms that lay perfectly flat on cover slip.
- myriophyllum
Thanks a lot. Right, flatness was a problem.
The designer Henry Siedentopf of Zeiss wrote about a maximum nA of 1.5 (Lit.: Kolloid-Zeitschrift; December 1925, Volume 37, Issue 6, pp 327–335; Ueber Neuerungen zur Dunkelfeldbeleuchtung; Henry Siedentopf).
The glas of the slide wouldn't give more anyway...
I think, the 14mm stop makes a 1.45 - 1.5 nA range, quite low light.
I don't have a nA 1.2 objective here, but it should work this way with an Ultrakondensor or other 1.2 - 1.4 condensers. The diameter of the stop needs to be determined very exactly...
Greetings
Jens
Greetings
Jens
zzffnn wrote:Very nice work, Jens!
- carlos.uruguay
- myriophyllum